9.ERROR CODE
103
+LOAD (A/D converter scale over)
Magnitude of signal from the load cell exceeds the F159’s adjustable range.
Check if the load cell output is properly within the range of Span calibration, or
if any of the cables are broken. This error can also occur when the terminal block
connections are all open.
-LOAD (A/D converter scale under)
Magnitude of signal from the load cell falls below the F159’s adjustable range.
Check if the load cell output is properly within the range of Span calibration, or
if any of the cables are broken. This error can also occur when the terminal block
connections are all open.
OFL1 (Net weight > Net over)
Net weight is larger than the set value for Net over. To restore normal weight
display, reduce the signal from the load cell until OFL1 bit (10th bit of n+7 CH)
turns off, or modify the set value for Net over.
OFL2 (Gross weight > Ca 9 scale division)
Gross weight is larger than Ca 9 scale division. To restore normal weight
display, reduce the signal from the load cell until OFL2 bit (11th bit of n+7 CH)
turns off
OFL3 (Gross weight > Gross over)
Gross weight is larger than the set value for Gross over. To restore normal weight
display, reduce the signal from the load cell until OFL3 bit (12th bit of n+7 CH)
turns off, or modify the set value for Gross over.
Maximum load (Capacity) is one of the most fundamental parameters for
F159 to perform properly as a metering instrument. Modification of Capacity
entails re-calibration of the system. Care should be taken to prevent
accidental modification of this value when you are correcting OFL2 error.
Request