PS-96 AC INPUT SECTION
PAGE
|
27
Important Note: Always re-install the terminal guard to protect users against inadvertent contact
with hazardous DC voltages on the connection lugs and DC output terminals.
1
Follow your facility’s Lockout/Tagout procedures and verify that input AC power to the DC
power supply is physically locked out or disconnected.
2
Connect the safety ground (earth) wire, typically green, to the input terminal labeled “GND”.
3
Connect the first AC three-phase hot wire, typically black, to the input terminal labeled “L1/L”.
4
Connect the second three-phase hot wire, typically red, to the input terminal labeled “L2/L”.
5
Connect a jumper wire (the same wire gauge as the other AC wires) between the terminals
labeled “L2/L” and “L2/N”.
6
Connect the third AC hot wire, typically blue, to the input terminal labeled “L3/N”.
Single-phase input power requirements for NOVANTA PS-96 DC power supply are 208–264 VAC, 1Ø
(single phase), 20 A maximum, 47–63 Hz. Table below provides recommendations for single-phase input
wiring and fusing.
Single-phase operation
Table 2-5 Single-phase AC electrical recommendations.
Parameter
Recommendation
Input voltage
208–264 VAC, 1Ø
Input current, max.
20 Amperes
Wire gauge
10 AWG
Fuse/circuit breaker
30 Amps
There are two possible circuit configurations for single-phase connections: (1) one circuit where the
voltage between hot conductors measures 208–264 VAC (recommended), or (2) an alternate case using
two individual circuits where the voltage between the hot and neutral conductors from each individual
circuit measures 120–264 VAC.
To connect single-phase AC input power to the PS-96 power supply, refer
back to prior figures and perform the following steps:
Important Note: Because AC input connections and requirements vary from facility to facility,
customers must provide the AC power cable or wiring.
One single-phase circuit (208–264 VAC)
1
Follow your facility’s Lockout/Tagout procedures and verify that input AC power to the DC power
supply is physically locked out or disconnected.
2
Connect the safety ground (earth) wire, typically green, to the input terminal labeled “GND”.
3
Connect one AC hot wire, typically black, to the input terminal labeled “L1/L”.
Содержание f201
Страница 1: ...ENGINEERED BY SYNRAD f201 Laser User Manual...
Страница 15: ...INVENTORY PAGE 15 Inventory F201 Lasers Figure 2 3 f201 shipping box contents...
Страница 36: ...F201 LABEL LOCATIONS PAGE 36 f201 label locations Figure 3 1 100 hazard label locations...
Страница 37: ...F201 LABEL LOCATIONS CONTINUED PAGE 37 f201 label locations continued Figure 3 2 F 200 hazard label locations...
Страница 38: ...F201 LABEL LOCATIONS CONTINUED PAGE 38 f201 label locations continued Figure 3 3 F 201 hazard label locations...
Страница 44: ...COMPLIANCE PAGE 44 Figure 3 4 f201 Declaration Document...
Страница 45: ...COMPLIANCE PAGE 45 Figure 3 5 continued F201 Declaration Document...
Страница 76: ...INTEGRATING F201 SAFETY FEATURES PAGE 76 Figure 4 15 F201 packaging diagram...
Страница 77: ...F201 GENERAL SPECIFICATIONS PAGE 77 f201 general specifications Table 4 8 f201 general specifications...
Страница 78: ...F201 GENERAL SPECIFICATIONS PAGE 78 Table 4 9 f201 general specifications continued...
Страница 89: ...STATUS LEDS PAGE 89 Table 5 6 Over Temperature fault Table 5 7 Shutter closed condition Table 5 8 VSWR fault...
Страница 104: ...INDEX PAGE 1 This page is intentionally left blank...