COOLING CONNECTIONS
PAGE
|
21
Important Note: DO NOT use de-ionized (DI) water as a coolant. DI water is unusually corrosive
and is not recommended for mixed material cooling systems.
Setting coolant temperature
Choosing the correct coolant temperature is important to the proper operation and longevity of your laser.
When coolant temperature is lower than the dew point (the temperature at which moisture condenses out
of the surrounding air), condensation forms inside the laser housing leading to failure of laser electronics
as well as damage to optical surfaces.
The greatest risk of condensation damage occurs when the laser is in a high heat/high humidity environ-
ment and the chiller’s coolant temperature is colder than the dew point of the surrounding air or when the
system is shut down, but coolant continues to flow through the laser for extended periods of time.
The chiller’s temperature setpoint must always be set above the dew point temperature. In cases where
this is not possible within the specified coolant temperature range of 18 °C to 22 °C (64 °F to 72 °F), then
the following steps MUST be taken to reduce the risk of condensation damage.
•
Use the Gas Purge port (f201) to introduce nitrogen or dry, filtered air into the laser housing.
•
Air-condition or dehumidify the room or the enclosure containing the laser.
•
Increase coolant flow by an additional 3.8 lpm (1.0 GPM). Do not exceed a coolant pressure of 414
kPa (60 PSI).
Important Note: To reduce problems associated with condensation, F201 laser f201asers can
operate up to 28 °C. However, these elevated cooling temperatures may result in decreased laser
performance and/ or reduced laser lifetime.
Содержание f201
Страница 1: ...ENGINEERED BY SYNRAD f201 Laser User Manual...
Страница 15: ...INVENTORY PAGE 15 Inventory F201 Lasers Figure 2 3 f201 shipping box contents...
Страница 36: ...F201 LABEL LOCATIONS PAGE 36 f201 label locations Figure 3 1 100 hazard label locations...
Страница 37: ...F201 LABEL LOCATIONS CONTINUED PAGE 37 f201 label locations continued Figure 3 2 F 200 hazard label locations...
Страница 38: ...F201 LABEL LOCATIONS CONTINUED PAGE 38 f201 label locations continued Figure 3 3 F 201 hazard label locations...
Страница 44: ...COMPLIANCE PAGE 44 Figure 3 4 f201 Declaration Document...
Страница 45: ...COMPLIANCE PAGE 45 Figure 3 5 continued F201 Declaration Document...
Страница 76: ...INTEGRATING F201 SAFETY FEATURES PAGE 76 Figure 4 15 F201 packaging diagram...
Страница 77: ...F201 GENERAL SPECIFICATIONS PAGE 77 f201 general specifications Table 4 8 f201 general specifications...
Страница 78: ...F201 GENERAL SPECIFICATIONS PAGE 78 Table 4 9 f201 general specifications continued...
Страница 89: ...STATUS LEDS PAGE 89 Table 5 6 Over Temperature fault Table 5 7 Shutter closed condition Table 5 8 VSWR fault...
Страница 104: ...INDEX PAGE 1 This page is intentionally left blank...