background image

 

 

QSG_F670_363.00_EN_01_2021-02-09

 

 

38/42 

 

Additional functions 

 
 

Label

 

Function

 

Description

 

24V OUTPUTS 

Auxiliaries output 24 

Available outputs for external automatism 

ABSOLUTE 

Absolute 

Display the absolute value of the results 

ATF 

ATF time 

Absorb the important leak variations at the defined time 

ATR0 

ATR1 

/

ATR2 

ATR3

 

ATR 0 - 3 

Specific filters on leak measurement 

AUTO CONNECT 

Automatic connector  Function to manage automatic jigs 

BUZZER 

Buzzer 

Buzzer activation configuration 

BYPASS 

Bypass function 

External fast filling valve management 

CODE READER 

Bar code reader 

Bar code configuration 

CUT OFF 

Cut off 

All the measurements that are lower than the configured rate 
have the value 0 

DISP. OPT. 

Display option 

Display of an additional information on a second line 

DISPLAY MODE 

Display Mode 

Leak measurement resolution 

END OF CYCLE 

End of cycle 

Several automatism case depending on fail part 
management 

FILL MODE 

Fill types 

Special filling methods 

FILTER 

Filtering 

Stabilize the measurement values 

FLOW LEVEL 

Flow level 

Add a minimum fail parameter 

INDIRECT 

Indirect mode 

Specific test mode: test part under bell 

MINI-VALVE 

Mini valve 

Access to highest time resolution (fast test) and auto zero 
time 

NAME 

Name 

Program customization 

NO NEGATIVE 

No Negative 

Replace negative value per 0 

N TESTS 

N tests 

Repeats the test when the results get close to the reject level 

PEAK HOLD 

Peak hold 

Give as result, the highest flow during the test time 

PR:SEQUENCE 

Sequencing 

Allowed program automatic sequencing 

PRE-FILL

 

Pre-fill types

 

Special filling methods

 

PRESS. CORR.

 

Pressure correction

 

Calculate leak measurement at a specific pressure value

 

PRESSURE DROP

 

Pressure drop

 

Pressure drop mode function in the Desensitized mode 

REF. VOLUME 

Reference volume 

Adjust the reference volume value with flow units only 

REWORK LIMIT

 

Rework limits

 

Additional levels for specific reworkable parts

 

Sd Part / SD PART 2 /  
SD PART 3 

Sealed part 

Several optional ways to test sealed parts 

SEALED DIFF 

Sealed differential 

Allows testing small parts volume difference between test 
and reference volumes 

SIGN

 

Sign

 

Return opposite result

 

STAMPING

 

Stamp

 

Pneumatic or electric output to identify the part

 

STD CONDITIONS

 

Standard conditions

 

Standard conditions correction with parameters

 

SYNC. TEST

 

Synchro test

 

A programmable input allows to pass from Stabilization to 
Test phase 

T+R TEST 

Test and reference 
test 

Display which part has failed during a test with two parts 

TEMP.CORR. 1 

Temperature 
correction 

Software temperature test part correction 

TEMP.CORR. 2 

Temperature 
correction 

Hardware temperature test part correction 

TEST CHECK 

Calibration check 

Calibration check by adding a volume at the end of test time 

Summary of Contents for F670

Page 1: ...ATEQ F670 Quick Start Guide...

Page 2: ...Connectors on the back panel with all options 10 Power supply connectors 12 Digital links 12 Analog outputs 15 Digital inputs outputs 16 Pneumatic connectors 19 Pneumatics configuration 22 User interf...

Page 3: ...ateq com cn T 86 21 6763 9508 F 86 21 6763 9528 ATEQ SYSTEMS ANALYSIS TAIWAN CO LTD NO 3 LAN 223 San Jia Dong Street 40642 TAICHUNG TAIWAN ateqtaiwan ateq com tw ateq com tw T 886 4 2437 5278 F 886 4...

Page 4: ...d the user manual before using the device All electrical connections to the device must be equipped with safety systems fuses circuit breakers etc adapted to the needs and in accordance with the appli...

Page 5: ...tions impurities must be prevented from being drawn into its internal components For this purpose we strongly recommend that a suitable airtight filter is installed between the part under test and the...

Page 6: ...QSG_F670_363 00_EN_01_2021 02 09 6 42 Preamble ATEQ F670 AN UNIVERSAL LEAK TESTER ATEQ F670 is a leak detector that tests the airtightness of parts ATEQ F670 can memorise 128 different test programs...

Page 7: ...by a cap 1 Device 2 Pressure sensor 3 Part under test 4 Differential pressure sensor 5 Reference part Desensitized test This mode is used for the measurement of large leaks when the reject level requ...

Page 8: ...ent cycle is made of 4 main phases fill stabilization test dumping A Pressure auto zero additional phase 0 can be placed at the start or at the end of a cycle depending on the requirement of the opera...

Page 9: ...t panel 1 Device name 2 Serial number 3 Display 4 Parameters lock key option 5 RS232 connector option 6 Mechanical regulator option 7 Cycle keys 8 Navigation keys 9 USB connectors 10 Mechanical regula...

Page 10: ...QSG_F670_363 00_EN_01_2021 02 09 10 42 CONNECTORS ON THE BACK PANEL WITH ALL OPTIONS...

Page 11: ...r option 14 Test part exhaust also for test pressure sensor auto zero 15 ATEQ Air supply energy information 16 Part number Serial number 17 P1 External back pressure connector option 18 T Test output...

Page 12: ...KS PC USB connectors on front face USB connectors can be used for connecting miscellaneous compatible USB devices The USB connectors are located under the rubber cover 1 see figure 1 Rubber cover 2 US...

Page 13: ...icenet connectors J5 J6 option M12 type connector 5 pins male connector J5 Devicenet input For connection to others ATEQ devices Pin number Signal 1 Not used 2 RXD data input 3 TXD data input 4 Not us...

Page 14: ...connector 4 pins female connector J5 J6 Ethernet connector J5 J6 option Standard connection Ethernet TCP IP protocol One of these network protocols is available Ethernet IP Profinet Ethercat J5 Input...

Page 15: ...M12 type connector 8 pins female connector J1 option Pin number Description 1 Ground pressure 2 0 10V DC pressure 3 Ground Pressure Diff 4 0 10V DC Pressure Diff 5 Signal contact event option 6 Ground...

Page 16: ...tion Characteristics Inputs Activation 24 V DC Outputs Dry contacts 60 V AC DC max 200 mA max Pin number Inputs outputs Description 1 Input 1 RESET 2 24 V DC Common 3 Input 2 START 4 24 V DC Common 5...

Page 17: ...the program number Pin number Inputs outputs Description 1 Input 8 Program selection from 33 to 64 programmable input 2 Input 9 Program selection from 65 to 128 programmable input Program number J9 J...

Page 18: ...V DC Common outputs 1 2 3 2 Output 1 Open collector 3 Output 2 Open collector 4 Output 3 Open collector 5 24 V DC Common outputs 4 5 6 6 Output 4 Open collector 7 Output 5 Open collector 8 Output 6 Op...

Page 19: ...ed by ATEQ Refer to Good practices and safety instructions section A specific filter may be necessary The air is supplied via the filter located on the back panel of the device Metal air filter The me...

Page 20: ...ons must be air tight Test and reference outputs 1 Test connector 2 Reference connector 3 Not used 4 Exhaust output 5 Pressurization output Metallic fitting available for test 1 and reference 2 connec...

Page 21: ...configuration 3 Optional fitting according to special configuration Pneumatic output 0 6 MPa 87 PSI option A and B automatic connectors option These connectors are used to drive pneumatic caps on the...

Page 22: ...at a pressure range from 0 to 500 kPa 72 5 PSI Connection Option description Air supply 1 Connection of the air supply to the filter input 0 6 MPa 87 PSI Vacuum to 2 Connection of the vacuum option 3...

Page 23: ...Connection of the regulator air supply to the filter input 1 2 MPa 174 PSI Air supply to 2 Connection of the air supply 0 6 MPa 87 PSI Vacuum to 3 Connection to the vacuum option 4 to 5 Connection of...

Page 24: ...tor air supply to 2 Connection of the regulator air supply 2 MPa 290 PSI 3 to ATM Output to be left to the atmosphere exhaust of electronic regulator 7 to 4 Connection of the Test output to the Y valv...

Page 25: ...ter input 0 6 MPa 87 PSI Regulator air supply to 2 Connection of the regulator air supply 2 MPa 290 PSI 3 to ATM Output to be left to the atmosphere exhaust of electronic regulator 7 to 4 Connection o...

Page 26: ...y to the filter input 1 2 MPa 174 PSI Air supply to 2 Connection of the air supply 0 6 MPa 87 PSI Vacuum to 3 Connection to the vacuum option 4 to ATM Output to be left to the atmosphere to exhaust th...

Page 27: ...ir blowing and temperatures variations Connection Option description Air supply to 1 Connection of the air supply to the filter input 0 6 MPa 87 PSI Vacuum to 2 Connection to the vacuum option 3 to 4...

Page 28: ...r supply to the filter input 0 6 MPa 87 PSI Vacuum to 2 Connection to the vacuum option 3 to 4 Connection of the reference output to the reference part 5 to 6 Connection of the V1 output to the refere...

Page 29: ...ction Option description Air supply to 1 Connection of the air supply to the filter input 0 6 MPa 87 PSI Vacuum to 2 Connection to the vacuum option 3 to 4 Connection of the reference output to the re...

Page 30: ...he front panel 1 Display 2 Cycle keys 3 Navigation keys KEYS Cycle keys The cycle keys are used to start and to stop a measurement cycle Key Name Function Start On the Program screen starts a measurem...

Page 31: ...gram screen Use the Program screen to select a test program 1 Current program name here NAME 2 Current program number here 001 3 Test type here LEAK TEST Access at startup of the device or by pressing...

Page 32: ...asurement unit 7 when the standard conditions function is validated Refer to the Reference Manual The MAIN MENU screen The MAIN MENU screen gives access to different sections for managing the device a...

Page 33: ...h last program used on screen PREPARING A PROGRAM Use this procedure to configure a new test program On the MAIN MENU screen ACCESSING THE PARAMETERS 1 Select PARAMETERS using the up down keys and the...

Page 34: ...icon is displayed at the bottom of the screen you must insert the USB unlocking device or enter a password before modifying a parameter On the PARAMETERS screen of the program see Preparing a program...

Page 35: ...screen STARTING A MEASUREMENT CYCLE 1 Press Start The cycle phases of the program are successively displayed FILL STABILIZATION TEST DUMP At the end of the cycle the results are displayed and READY a...

Page 36: ...se the part with an infinite fill time Piezo auto zero Piezo auto zero Auto zero cycle on the piezo sensor AUTO VOL Automation volume Volume calculation for automatic program selection Label Special c...

Page 37: ...for the test part Above this limit the part is considered as defective Ref FAIL Reference fail Reference part reject level FUNCTIONS Functions Access to additional functions Some parameters are displa...

Page 38: ...E No Negative Replace negative value per 0 N TESTS N tests Repeats the test when the results get close to the reject level PEAK HOLD Peak hold Give as result the highest flow during the test time PR S...

Page 39: ...process option Press UNIT Pressure unit by default for the new programs DUMP LEVEL Configuration of the minimum dump level pressure LINE P MIN Minimum level for checking line pressure option BLOW MODE...

Page 40: ...al cycles to carry out device internal tests RESET PARA Parameters reset Reset to factory configuration FIELDBUS Fieldbus parameters Fieldbus information RESULTS menu In this section manage measuremen...

Page 41: ...is section describes save and restore parameters on an external USB device Label Description Save parameters Save parameters on an external USB memory device for restoring later Restore parameters Res...

Page 42: ...I Pneumatic air supply 0 to 0 5 MPa range Air supply 0 6 MPa 15 Pneumatic air supply 0 6 to 1 MPa range Regulator input 1 2 MPa 10 Valves supply 0 6 MPa 15 Pneumatic air supply 1 1 to 2 MPa range Regu...

Reviews: