4.CALIBRATION
45
When a calibration error is displayed, take measures according to the error message, and do zero
calibration again.
CC-Link Console
1 - 2
…
The initial tare is above the Zero adjustment range.
Remove any excess load from load cell or scale. If cErr2 is still
displayed, connect a resistor bEX and -SIG load cell
connections. This should shift the Zero point. Do Zero Calibra-
tion again.
For the relationship between the resistance connected and input
signals, see the next page.
1 - 3
…
The initial tare is minus.
Check that load cell is mounted in the correct direction; check
that load is being applied to the load cell in the correct direction;
check that the +SIG and -SIG lines are properly connected.
If cErr3 is still displayed, connect a resistor between -EX and -
SIG load cell connections. This should shift the Zero point. Do
Zero Calibration again.
For the relationship between the resistance connected and input
signals, see the next page.
1 - 9
…
The weight value is not stable.
Adjust the parameters for motion detection, check that the
STAB status is “ON”, and then do zero calibration again.
* Each CC-Link error code consists of:
Error code
Assistance code
-
Содержание F156
Страница 1: ...F156 WEIGHING CONTROLLER 07 JUN 2012 REV 1 40 OPERATION MANUAL CC Link Ver 1 10...
Страница 115: ...7 REMOTE CONSOLE UNIT 106 Monitor change Change the display with the key...
Страница 120: ...7 REMOTE CONSOLE UNIT 111 Monitor change Change the display with the keys 0 to 9 1 2 3 4 5 0...
Страница 155: ...13 DIMENSIONS 146 13 DIMENSIONS...
Страница 156: ...14 BLOCK DIAGRAM 147 14 BLOCK DIAGRAM...
Страница 160: ...Unipulse Corporation 9 11 Nihonbashi Hisamatsucho Chuo ku Tokyo 103 0005 Tel 81 3 3639 6120 Fax 81 3 3639 6130...