Chapter
11
■
Sub numerical display section
-LoAd
Beyond the signal input
range -3.2mV/V
The input electric signal greatly exceeds the signal
input range. This also occurs when an excessive load
is applied to the sensor or the input terminal is opened
due to a break in the cabling, etc.
LoAd
Beyond the signal input
range +3.2mV/V
ALA
Exceeds the alarm HI limit
This occurs when a larger load than expected is
applied to the sensor or calibration is performed below
the measurement range. Check the calibration and
measuring object.
-ALA
Below the alarm LO limit
oFL1
Below 19999
This occurs when a larger load than expected is
applied to the sensor, or calibration is performed below
the measurement range. Check the calibration and
measuring object.
oFL2
Exceeds -19999
cErr2
Beyond the zero calibration
range in the positive
direction
The signal input range is greatly exceeded when zero
calibration is performed. Check for an excessive load
applied to the sensor and for breaks.
cErr3
Beyond the zero calibration
range in the negative
direction
cErr4
Beyond the rated output
value in the positive
direction/negative direction.
Beyond the rated capacity
value.
The signal input range is greatly exceeded when
calibration is performed. In the case of equivalent input
calibration, check the rated output value, and in the
case of actual load calibration, check for an excessive
load applied to the sensor and for breaks.
Also, Check the setting of rated capacity value (display
value) sets it exceeding "±19999".
cErr5
Rated capacity value = 0
Check the setting of rated capacity value (display
value) for being "0" when calibration is performed.
cErr6
Rated output value
0.000mV/V at the time of
equivalent input calibration
Rated output value in the
range of -0.005 to
+0.005mV/V at the time of
actual load calibration
Check the rated output value for being 0.000mV/V
when equivalent input calibration is performed, and the
load for being too light in weight when actual load
calibration is performed.
dZ Err
Digital Zero limit error
When the range of zero point correction (deviation from
the zero calibration registered point) by Digital Zero or
Zero Tracking exceeds the setting value of Digital Zero
limit, Digital Zero limit error occurs. Perform Digital
Zero, etc., in a proper range.
CALZE
Zero calibration in progress
These are not errors. They indicate that each
calibration is in execution.
CALSP
Actual load calibration in
progress
CALEq
Equivalent input calibration
in progress