Measurement range
A AC/DC
0.30...59.99 A
60.0...599.9 A
600...1000 A RMS/DC
or 1500 A PEAK
Accuracy
2% rdg
±
2 0.2 A
2% rdg
±
2 counts
(with corrected zero)
Resolution
10 mA
100 mA
1 A
NB : Residual current on zero :
- on DC : a function of the temperature and the residual magnetism. It can be corrected by the DC zero function
- on AC :
≤
200 mA
PEAK mode
An additional error of 0.5 % on peak value, is to be added to the error specified in RMS or DC mode.
Residual current on peak value : 0.3 A max (corrected zero).
RECORD PEAK mode
An additional error of 1 % on peak value, is to be added to the error specified in RMS or DC mode.
Residual current on peak value: 0.5 A max (corrected zero).
5.3
DC AND AC VOLTAGES
Particular reference conditions :
- On AC : sinusoidal signal
- On DC : AC component
≤
0.1% of the DC signal
!
Limits on voltage measurements
- Operating frequency
: 10 Hz to 5 kHz
- Permitted permanent overload : 1.5 kV PEAK
NB : Affichage de
!
: - steady above 600 V RMS
- flashing above 1500 V PEAK or DC
Measurement range
V AC/DC
0.05...59.99 V
60.0...599.9 V
600 V RMS/DC
or 600...1500 V peak
(1)
Accuracy
1% rdg
±
10 counts
1% rdg
±
2 counts
Resolution
10 mV
100 mV
1 V
Input impedance
1 M
Ω
(1)
May be used up to 1500 V PEAK on condition that the corresponding rms value is < 600 V (IEC 1010).
Mode PEAK
An additional error of 0.5 % on peak value, is to be added to the error specified in RMS or DC mode.
Residual voltage on peak value: 0.3 V max.
RECORD PEAK mode
An additional error of 1 % on peak value, is to be added to the error specified in RMS or DC mode.
Residual voltage on PEAK value: 0.3 V max.
5.4
CREST FACTOR (CF)
Particular reference conditions
- Max peak value 1500 V or 1500 A
- Min value of signals 5 V and 5 A
- Operating frequency 40 Hz to 70 Hz