Specific characteristics in PEAK mode
(from 10 Hz to 1 kHz in AC):
•
Uncertainties: add ± (1.5% L+0.5A) to the values in the tables above.
•
PEAK capture time: 1ms min. to 1.5ms max.
4.2.8
Calculation of the crest factor (CF)
Measurement range
1.00 – 3.50
3.51 – 5.99
6.00 – 10.00
Specified
measurement range
(from 5V or 5A)
0 to 100% of the measurement range
Uncertainties (zero
corrected in A DC)
± (2% R + 2 pt)
± (5% R +2 pt) ± (10% R + 2 pt)
Resolution
0,01
Remark :
Peak values limited to 1500V or 1500A
Uncertainties guarantied until 400 Hz
4.2.9
Calculation of the RIPPLE in DC
Measurement range
0,01% - 99,99%
100,0% - 1000%
Specified measurement
range (from 3 A DC and
2 V DC)
2 to 100% of the
measurement range
0 to 100% of the
measurement
range
Uncertainties
± (5% R +10 pt)
Resolution
0,01
0,1
Remark :
If one of the terms for the calculation of the RIPPLE is displayed as "OL",
or forced to zero, the RIPPLE displayed is an indeterminate value, "----".
4.2.10 Continuity measurement
Measurement range
0.0
Ω to 999.9
Ω
Open-circuit voltage
≤
3.6V
Measurement current
550 µA
Uncertainties
± (1% R +5 pt)
Buzzer triggering threshold
40
Ω