7
M10*55+ flat washer 10+spring
washer 10
Hexagon wrench
1pc
User Manual
1pc
Envelope(Behind the cabinet)
Inspection report
1pc
Certification of quality
1pc
Warranty Card
1pc
1.3 Installation Conditions and Operating Environment
Laminar flow cabinet shall be placed in a protection area, workspace can not be right opposite to the
door or window, and away from the outlet of air conditioning, prevent from the ventilation system,
air conditioning, door, window, and personnel movement caused by the airflow impact on Laminar
flow cabinet.
At least 300mm gap must be kept in the side and back side of the Laminar Air Flow for clean
operating and for inspection.
Working environment:
(1) Only applicable for indoor operation
(2) Ambient temperature: 15° C
~
35° C
(3) Relative Humidity: ≤75%
(4) Atmospheric pressure range: 70KPa
~
106KPa
(5) Electrical parameters: Adequate power supply to the laminar flow cabinet(See 2.1.4 Technical
Parameters)
(6) Power supply need to be grounded; (Judging method: test the live wire and the neutral wire of
the socket with multimeter. The voltage between live and ground should equal to the voltage of local
electrical grid, and the voltage between neutral and ground should equal to 0. Otherwise the power
supply is not grounded correctly)