Introduction 

The electromagnetic flowmeter uses Faraday’s Law of 
electromagnetic induction to measure the process flow. 
The  device  consists  of  two  units:  a  detector,  through 
which  the  fluid  to  be  measured  flows  and  in  which 
low-level  signals  proportional  to  flow  rates  are 
obtained;  and  a  converter,  which  supplies  excitation 
current to the detector, and amplifies the signals from 
the detector and then processes and converts the signals 
into  the  4–20mAdc  current  signal  or  communication 
signal.  The  meter  features  a  Mount-  Anywhere 
magnetic  field  distribution  technology,  the  meter  is 
highly  immune  to  upstream  flow  disturbances. 
Combined  with  a  multi-functional  converter  LF620 
(combined  type)  or  LF622  (separate  type)  equipped 
with  its  patented  Noise-Sentry  original  noise- 
suppression  circuit  and  advanced  algorithms.  The 
LF410  has  a  very  high  tolerance  to  noise,  giving  the 
unit  a  very  stable  output  even  for  slurry  fluid 
measurement. IR (Infrared) switches enable parameter 
setting  of  the  converter  without  removing  the  cover. 
Flow direction can be set in either way, and its unique 
128 x 128 dot matrix LCD display allows the LCD to 
be  rotated  electronically  to  90,  180  and  270  degrees 
without opening the cover.  

*1: HART protocol (Highway Addressable Remote 

Transducer) is a communication protocol for industrial 
sensors recommended by the HCF

 

(HART

 

Communication

 

Foundation).

  

 

** DevComm2000 Smart Device Communicator available for 
performing HART device configurations on PC or laptop is 
available through the supplier . 

*2: PROFIBUS is the communication protocol for  factory 

and process automation that the PROFIBUS 
Organization recommends. Instead of analog control 
with a conventional analog signal (4-20mA), it is 
fieldbus which digitizes all signals. Flowmeters support 
PROFIBUS-PA. 

Modbus is the communication protocol that Modicon Inc. 
developed.  Physical layer is RS485.

 

 

Power   
supply 

Converter 

I/O 

Detector 

Combined type 

LF410/LF620 

  LF414/LF620F 

Converter 

Power 
supply 

Signal cable 

Detector 

Separate type 
LF410/LF622 

  LF414/LF622F 

Terminal box 

I/O 

 

Figure1.  Configuration 

 

 

 

LF410/LF620   

LF410   

LF622 

LF414/LF620F 

LF414                    LF622F 

Figure2.  LF410 Mount-Anywhere series 

Flowmeters 

 

                     

 

 

Specifications 

 Overall Specifications 

Measurement range in terms of flow velocity: 

1.0

 

ft/s to 32.8

 

ft/s (0.3

 

m/s to 10

 

m/s). 

0.3

 

ft/s to 1.0

 

ft/s (0.1

 

m/s to

 

0.3

 

m/s)  

range is available optionally. 

Accuracy: 

±

±

±

±

0.2 % of Rate

* 

* This pulse output error result is established under standard 

operating conditions at Toshiba's flow calibration facility, 
Fuchu Japan. (NIST Traceable).  

* Individual meter measurement error may vary up to ±0.5% of 

Rate at 1.64ft/s (0.5m/s) or more and ±0.3% of rate ±0.039 
inch/s (1mm/s) at 1.64 ft/s (0.5 m/s) or less. 

* Current output: plus ± 8µA (0.05% of span.) 

* Refer to individual calibration data for each individual meter's 

measurement error. 

Fluid conductivity: 5µS/cm minimum 

Fluid temperature:  

14 to 356 

°

F (–10 to +180

°

C):Separate type for 

  ceramic tube. 

14  to  248

°

F  (–10  to  +120

°

C):  Combine  type  for 

ceramic tube 

 

Ambient temperature: -4 to 140 °F (-20  to  + 60 °C) 

Field Intelligent Device – Mount-Anywhere Series - Wafer 

Electromagnetic Flowmeter 

 

LF410 /LF620 
LF410 /LF622 

1/2" to 4" (15 to 100 mm) 

Certification number 

Z01207 

 

 

 

Summary of Contents for F414

Page 1: ...op is available through the supplier 2 PROFIBUS is the communication protocol for factory and process automation that the PROFIBUS Organization recommends Instead of analog control with a conventional analog signal 4 20mA it is fieldbus which digitizes all signals Flowmeters support PROFIBUS PA Modbus is the communication protocol that Modicon Inc developed Physical layer is RS485 Power supply Con...

Page 2: ...the manufacturer each detector unit is tested by following test condition Pressure 600 psi Time 5 minutes Connection flange standards ASME B16 5 class 150 ASME B16 5 class 300 EN1092 1 PN10 EN1092 1 PN16 JIS B 2220 10K JIS B 2220 16K JIS B 2220 20K Principal materials Case 1 to 4 25 to 100mm stainless steel 1 2 15mm carbon steel Linings 1 2 to 4 15 to 100mm Ceramic tube Electrodes Type Super smoot...

Page 3: ...lowing functions can be assigned to DO1 and or DO2 Pulse output available only for DO1 DO2 Pulse rate Max 10kHz 10 000pps DO1 Max 100Hz 100pps DO2 Over 1kpps auto setting Pulse width 0 3 to 500ms but less than half of the period for 100 flow rate Note The same and simultaneous pulse is not available between DO1 and DO2 Multi range selection outputs Note 1 High High high Low and or Low low alarm ou...

Page 4: ...the meter from lightning and improve personnel safety Case Aluminum alloy equal to IP67 Coating Acrylic resin baked coating pearl gray colored Cable connection port Cable glands LF620 and LF622 without cFMus Approval Provided as standard OD of cableφ11 13mm Material Nylon 66 G PF 1 2 male threads LF620F and LF622F with cFMus Approval Not provided 1 2 14NPT male threads are required Applicable diam...

Page 5: ...PN 16 and JIS B 2220 10K dimensions Meter size mm L1 mm L2 mm D1 mm Weight kg 25 80 226 66 approx 4 Note 1 inch 25 4mm ASME B 16 5 class 150 and class300 dimensions Meter size inch L1 inch L2 inch D1 inch Weight lb 1 3 15 8 90 2 60 approx 9 Note 1 inch 25 4mm Figure 4 LF410 LF620 and LF414 LF620F flowmeters Meter size 1 25mm Meter size L1 1 2 15mm 3 03 inch 77mm 1 25mm 3 74 inch 95mm 1 1 2 40mm 4 ...

Page 6: ...ze inch L1 inch L2 inch D1 inch Weight lbs 1 1 2 3 94 9 80 3 35 approx 11 2 4 33 10 43 4 02 approx 14 3 4 33 11 46 5 00 approx 16 4 4 72 12 72 6 26 approx 20 The dimension of L1 changes when the material of grounding ring is chosen Pt Ir or Ta Meter size L1 1 2 15mm 3 03 inch 77mm 1 25mm 3 74 inch 95mm 1 1 2 40mm 4 53 inch 115mm 2 50mm 4 96 inch 126mm 3 80mm 4 96 inch 126mm 4 100mm 5 35 inch 136mm...

Page 7: ...dimensions Meter size inch L1 inch L2 inch D1 inch Weight lb 1 2 2 76 7 05 1 93 approx 7 EN 1092 1 PN16 and JIS B 2220 10K dimensions Meter size mm L1 mm L2 mm D1 mm Weight kg 15 70 179 49 approx 3 The dimension of L1 changes when the material of grounding ring is chosen Pt Ir or Ta Meter size L1 1 2 15mm 3 03 inch 77mm 1 25mm 3 74 inch 95mm 1 1 2 40mm 4 53 inch 15mm 2 50mm 4 96 inch 126mm 3 80mm ...

Page 8: ...mensions Meter size inch L1 inch L2 inch D1 inch Weight lb 1 3 15 6 61 2 60 approx 7 EN 1092 1 PN16 and JIS B 2220 10K dimensions Meter size mm L1 mm L2 mm D1 mm Weight kg 25 80 168 66 approx 3 The dimension of L1 changes when the material of grounding ring is chosen Pt Ir or Ta Meter size L1 1 2 15mm 3 03 inch 77mm 1 25mm 3 74 inch 95mm 1 1 2 40mm 4 53 inch 115mm 2 50mm 4 96 inch 126mm 3 80mm 4 9...

Page 9: ...3 8 15 4 02 approx 11 3 4 33 9 13 5 00 approx 13 4 4 72 10 39 6 26 approx 20 EN 1092 1 PN16 and JIS B 2220 10K dimensions Meter size mm L1 mm L2 mm D1 mm Weight kg 40 100 190 85 approx 4 50 110 207 102 approx 5 80 110 232 127 approx 6 100 120 264 159 approx 9 The dimension of L1 is changes when the material of grounding ring is chosen Pt Ir or Ta Meter size L1 1 2 15mm 3 03 inch 77mm 1 25mm 3 74 i...

Page 10: ... signal line A B G andthe excitation line E X Y are the connector joints Grounding with 100Ω or less ground resistance 1 Locate an external double pole power switch on the power line near the flowmeter within easy reach of operation Use the appropriate switch rating as shown below Switch rating 250Vac 6A or more In rush current 15A or more Figure 10 Combined type LF410 LF620 and LF414 LF620F flowm...

Page 11: ...conduit Power switch External double pole power switch Figure 11 Separate type LF410 LF622 and LF414 LF622F flowmeters wiring Diagram Table 2 LF620 LF620F LF622 and LF622F Converters Signal Table Symbol Description Cable L1 Power supply Power cable L2 GND Ground for arrester FG Frame ground DI Digital Input 20 30Vdc I O cable DO1 Digital Output 1 DO2 Digital Output 2 COM Signal Common for DI DO1 D...

Page 12: ...tion for cables be careful of allowable maximum bend diameter etc Basically do not install cables in a way cables are twisted or bent 5 Consider installing a PROFIBUS PA arrester in the communication path of PROFBUS PA so that the electromagnetic flowmeter will not be affected by lightning etc 6 The electromagnetic flowmeter is not equipped with terminating resistors Use the terminating resistor u...

Page 13: ...use an appropriate cover 1 Where unit operation panel is exposed to direct sunlight reflection of light onto window pane and diffused light reflection 2 Where smoke and steam may occur 3 Where exposed to direct snow ice or mud Ordering Information 1 When ordering the LF410 series flowmeters refer to Tables 5 and 6 Type Specification Codes An entry must be made for each of the columns in each of th...

Page 14: ...L stainless steel E Ti titanium H Hastelloy C Equivalent Z other Flow and calibration velocity range A 1 0 to 32 8 ft s standard range calibration B 1 0 to 32 8 ft s specified range calibration C 0 3 to 32 8 ft s specified range calibration Excitation and Signal Cables A not provided C other lengths provided Note 4 Coating A no coating B phthalic acid resin coating pearl gray colored Note 5 Size c...

Page 15: ...available Note 1 When PROFIBUS communication is provided current output 4 20mA and HART communication cannot be used When Modbus communication is provided digital output points 1 DO1 and digital output points DO2 digital input point 1 DI HART communication cannot be used Refer Table 7 for more details Note2 Select 110Vdc for test report inspected under the condition of 110Vdc Table7 Communication ...

Reviews: