Operation of the ThreadChecker 

Wire colors & pin-out.  

The wire colors and typical connection diagram are as follows. 

Teach sequence

 

Before the system is first used it must have the ‘teach’ sequence run. In order to do 

this you must have a representative threaded and non-threaded hole. The system can 

be easily calibrated to any new sensor designed for the ThreadChecker as well as for 

most hole/thread configurations with a simple calibration sequence. 

1. Connect the sensor and power source to the electronics. Allow the system to

warm up for a minimum of 5 minutes.

2. With the sensor in ‘air’ (i.e. not in a hole or near any conductive material), press

and hold the ‘Teach’ button until the Thread LED blinks fast then release (more

than 1 second but less than 10 seconds), the Thread LED should blink slowly.

3. Position the sensor tip in the threaded hole and momentarily press the ‘Teach’

button. The Status LED will blink fast.

4. Position the sensor in an unthreaded or bad hole and momentarily press the

‘Teach’ button. The LED should be solid red indicating no threads present. The

system is now calibrated and ready for use.

The

 switch point is 

set 

half way between these 2 points

. The

 teach 

process can 

also be performed in 

reverse order (bad then good) if desired. 

Summary of Contents for ThreadChecker

Page 1: ...Products Copyright 2019 PART NO 860510 001A Last Revised 2019 10 22 Measuring Memory Systems 217 Smith Street Middletown CT 06457 www kamansensors com or 860 632 4442 Page 1 of 12 ThreadCheckerTM Nonc...

Page 2: ...installed and operated per the manufacturer s recommendations conforms with the protection requirements of EC Council Directive 89 336 EEC on the approximation of the laws of the member states relatin...

Page 3: ...nsumption ThreadChecker Unit The ThreadChecker system consists of two subassemblies the sensor with integral cable and the signal conditioning electronics module with power and output cable The Thread...

Page 4: ...Adjustment and Calibration The ThreadChecker requires calibration to the users configuration before use NOTE solid red on the power LED means that it does not recognize the sensor or the sensor is di...

Page 5: ...ns with a simple calibration sequence 1 Connect the sensor and power source to the electronics Allow the system to warm up for a minimum of 5 minutes 2 With the sensor in air i e not in a hole or near...

Page 6: ...ty mode the switch is open when detecting no thread and closed to ground when it detects a thread or open air Inverting the Switch Polarity The polarity of this output can be inverted in either mode s...

Page 7: ...switch output is pulled to ground When the LED is on the output is open In standard Window comparator mode the status LED functionality is NOT affected by the limit inversion and will continue to oper...

Page 8: ...e sensor Ideally the sensor should be just slightly smaller than and centered in the untapped hole The surface of the sensor tip should be at mid depth Larger For holes that are much larger than the s...

Page 9: ...acquire a set of calibration data for all frequencies for the threaded hole The last step is to put the sensor in a non threaded hole and press the button again On release the system will acquire dat...

Page 10: ...tem not resolving thread no thread correctly Make sure sensor connector is screwed on tight before a teach Status LED never goes to red when there is no threaded hole Do a teach to make sure that it h...

Page 11: ...done using the two holes in the enclosure and an M 4 or 6 32 screw see below The DIN mount attachment can also be used to allow mounting on a DIN rail Positioning the electronics module so that it is...

Page 12: ...rent short circuit protected Impedance 50 ohms Voltage 0 10 VDC SWITCHED OUTPUT Opto Isolated Load Current 80 mA maximum AC or DC Load Voltage 30V rms 42 4V peak On Resistance 1 ohms minimum 8 ohms ma...

Reviews: