46 • User Guide
Guidance and manufacturer’s declaration – electromagnetic immunity
The F1 products are intended to be used in the electromagnetic environment specified below. The customer or the user of the F1 products
should assure that it is used in such an environment.
Immunity test
IEC 60601 test level
Compliance level
Electromagnetic environment - guidance
Electrostatic dis-
charge (ESD) IEC
61000-4-2
± 6 kV contact
± 8 kV air
± 2, ± 4, ± 6 kV contact
discharge
± 2, ± 4, ± 8 kV air discharge
± 2, ± 4, ± 6 kV indirect
contact discharge
Floors should be wood, concrete or ceramic tile. If
floors are covered with synthetic material, relative
humidity should be at least 30%.
Electrical fast transi-
ent/burst IEC 61000-
4-4
± 2 kV for power supply
lines
± 1 kV for input/output
lines
± 2 kV for power supply lines
Mains power quality should be that of a typical
commercial or hospital environment.
Surge IEC 61000-4-5 ± 1 kV line(s) to line(s)
± 2 kV line(s) to earth
± 0.5, 1 kV line(s) to line(s)
± 0.5, 1, 2 kV line(s) to earth
Mains power quality should be that of a typical
commercial or hospital environment.
Voltage dips, short
interruptions and
voltage variations on
power supply input
lines IEC 61000-4-11
<5 %
U
T
(>95 % dip in
U
T
) for 0,5 cycle
40 %
U
T
(60 % dip in
U
T
)
for 5 cycles
70 %
U
T
(30 % dip in
U
T
)
for 25 cycles
<5 %
U
T
(>95 % dip in
U
T
) for 5 s
<5 %
U
T
(100 % dip in
U
T
) for
0,5 cycle
40 %
U
T
(60 % dip in
U
T
) for 5
cycles
70 %
U
T
(30 % dip in
U
T
) for
25 cycles
<5 %
U
T
(100 % dip in
U
T
) for
5 s
Mains power quality should be that of a typical
commercial or hospital environment. If the user of
the Fimet F1 requires continued operation during
power mains interruptions, it is recommended to
power the Fimet F1 from an uninterruptible power
supply or a battery.
Guidance and manufacturer’s declaration – electromagnetic immunity
The F1 products are intended for use in the electromagnetic environment specified below. The customer or the user of the F1 products
should assure that it is used in such an environment.
Immunity test
IEC 60601 test level
Compliance level
Electromagnetic environment - guidance
Conducted RF IEC 61000-
4-6
Radiated RF IEC 61000-4-
3
3 Vrms 150 kHz to 80 MHz
3 V/m 80 MHz to 2,5 GHz
3 V
3 V/m
Portable and mobile RF communications equipment
should be used no closer to any part of the F1
product, including cables, than the recommended
separation distance calculated from the equation
applicable to the frequency of the transmitter.
Recommended separation distance
P
d
*
17
.
1
=
P
d
*
17
.
1
=
80 MHz to 800 MHz
P
d
*
33
.
2
=
800 MHz to 2,5 GHz
where P is the maximum output power rating of the
transmitter in watts (W) according to the transmitter
manufacturer and d is the recommended separation
distance in meters (m).
Field strengths from fixed RF transmitters, as
determined by an electromagnetic site survey (a),
should be less than the compliance level in each
frequency range (b).
Interference may occur in the vicinity of equipment
marked with the following symbol:
NOTE 1 At 80 MHz and 800 MHz, the higher frequency range applies.
NOTE 2 These guidelines may not apply in all situations. Electromagnetic propagation is affected by absorption and reflection from
structures, objects and people.
a) Field strengths from fixed transmitters, such as base stations for radio (cellular/cordless) telephones and land mobile radios, amateur
radio, AM and FM radio broadcast and TV broadcast cannot be predicted theoretically with accuracy. To assess the electromagnetic
environment due to fixed RF transmitters, an electromagnetic site survey should be considered. If the measured field strength in the use
location of the Fimet F1 exceeds the applicable RF compliance level above, the Fimet F1 should be observed to verify normal operation. If
abnormal performance is observed, additional measures may be necessary, such as re-orienting or relocating the Fimet F1.
b) Over the frequency range 150 kHz to 80 MHz, field strengths should be less than [V1] V/m.
Summary of Contents for F1
Page 1: ...Fimet F1 User Guide Version 1 1 1 6 2016 ...
Page 11: ...User Guide 11 F1 F1 Europa F1 Mondo F1 City F1 Prime F1 Ceiling F1 Cart F1 Cab ...
Page 12: ...12 User Guide 2 2 System Overview Container of the flushing liquid ...
Page 53: ...1 Appendix A Dimensions and space requirements 1 1 Dimensions F1 ...
Page 54: ...54 User Guide 1 2 Dimensions F1 Cab ...
Page 55: ...User Guide 55 1 3 Dimensions F1 Cart ...
Page 56: ...56 User Guide 1 4 Dimensions F1 Prime ...
Page 57: ...User Guide 57 ...
Page 58: ...58 User Guide 1 5 Dimensions F1 Side ...
Page 59: ...User Guide 59 ...