133
14
Error Message Display
133
Err
o
r Me
ssa
ge
Display
Chapter
14
■
Over scale display
■
Calibration error display
■
Error message display
Display
Error details
A/D converter Negative over (-3.9mV/V or less between ±SIGs)
A/D converter Positive over (+3.9mV/V or more between ±SIGs)
Display over range When below -19999 (indicated value = -19999)
Display over range When over 99999 (indicated value = 99999)
OVERLOAD error
When sensor input value exceeds alarm HI limit or alarm LO limit
(However, overload and so on are not monitored against sensor input values
when used in analog peak hold (peak hold selection = 1: analog, hold mode = 1:
peak hold), but is monitored against the indicated values (hold values during
hold).)
Display
Calibration
error No.
Error details
2
Zero calibration or span calibration cannot be performed
because the electronic signal input on the positive side
exceeds the input voltage range.
Check for cable disconnection and mis-wiring.
3
Zero calibration or span calibration cannot be performed
because the electronic signal input on the negative side
exceeds the input voltage range.
Check for cable disconnection and mis-wiring.
5
Sensor rated capacity value is set as "0".
Reset to an appropriate value.
6
The output of strain gauge type sensor has not
reached the span adjustment range (approx. 0.05 mV/V).
Check the load and confirm that the sensor has sufficient
performance for adequate output. Then, perform span
calibration again.
7
The strain gauge type sensor is output to the negative side.
Confirm that the + SIG and - SIG wires of the strain gauge
type sensor are not reversely connected.
Display
Error message details
Zero calibration
Actual load calibration, equivalent input calibration
14
Error Message Display