5.CALIBRATION
29
5. CALIBRATION
"Calibration" refers to an operation whereby matching between the F320 and a
strain gauge sensor is obtained. The F320 uses the two calibration methods as
described below.
◇
Equivalent Input Calibration
This approach uses no actual loads but key entry of the rated output value of the
strain gauge sensor (mV/V) and the rating value (value to be displayed).
This method is simple and employed when actual loads cannot be applied.
For example
Gain will be automatically decided by registering the values indicated as
follows:
for load:
2.001mV/V - 100.0kgf
for pressure:
2.002mV/V - 10.00kgf/cm2, and
for torque:
2.502mV/V - 15.00kgf•m.
◇
Actual Load Calibration
This approach provides calibration by applying an actual load to the strain gauge
sensor and inputting the actual load value. This calibration is without little errors
and more correct.
A data sheet will be attached to a strain gauge sensor you buy.
On the data seat, the following values are listed.
Capacity ................................... Load (unit: Kg, t, etc.)
Rated Output............................ Voltage (unit: mV/V)
Non-Linearity
,
Hysterisis
,
Input Resistance
,
Output Resistance
,
Zero Balance, etc.
The Capacity and the Rated Output are necessary values for the
equivalent input calibration. Input these two values to F320.
Содержание F320
Страница 1: ...F320 DIGITAL INDICATOR 17 May 2012 Rev 1 07 OPERATION MANUAL...
Страница 14: ...CONTENTS...
Страница 120: ...15 DIMENSIONS Standard 106 15 DIMENSIONS Standard Unit mm...
Страница 121: ...15 DIMENSIONS Standard 107 15 1 Dimensions When the BCO option is equipped Unit mm...
Страница 122: ...15 DIMENSIONS Standard 108 15 2 Dimensions When the DAV DAI option is equipped Unit mm...
Страница 129: ...Unipulse Corporation 9 11 Nihonbashi Hisamatsucho Chuo ku Tokyo 103 0005 Tel 81 3 3639 6120 Fax 81 3 3639 6130...