60
4.2.11 Resistance measurement
Measurement range
(1)
0.
0 Ω to
59.9
Ω
60.
0 Ω to
599.9
Ω
600 Ω to
5999
Ω
6.00 k
Ω to
59.99 k
Ω
Specified
measurement range
1 to 100% of the
measurement range
0 to 100% of the
measurement range
Uncertainties
± (1% R+10 pt)
± (1% R +5 pt)
Resolution
0.1
Ω
1
Ω
10
Ω
Open-circuit voltage
≤
3.6 V
Measurement current
550 µA
100 µA
10 µA
Note (1) - Above the maximum display value, the display unit indicates "OL".
The "-" and "+" signs are not managed.
Specific characteristics in MAX/MIN mode
:
•
Uncertainties: add 1% R to the values of the table above.
•
Capture time of the extrema: approximately 100ms.
4.2.12 Active DC power measurements
Measurement
range (2)
0 W to
9999W
10,00 kW
to
99,99kW
100,0 kW to
999,9 kW
1000 kW to
1500 kW (1)
Specified
measurement
range
1 to 100% of the
measurement range
0 to 100% of the measurement range
Uncertainties
(3)
until 1000A
± (2% R +10 pt)
from 1000A to 1500A
± (2.5% R +10 pt)
until 1000A
± (2% R +3 pt)
from 1000A to 1500A
± (2.5% R +3 pt)
Resolution
1W
10W
100W
1000W
Note 1 - Display of O.L above 1500kW in single-phase (1000V x 1500A).
Note 2 - Any applied voltage greater than 1000V causes the emission of an
intermittent alarm beep to report a dangerous overload.
Note 3 - The measurement result may be perturbed by an instability linked to the
current measurement (approximately 0.1A).
Example: for a power measurement made at 10A, the instability of the
measurement will be 0.1A/10A or 1%.