CALIBRATION PROCEDURE
33
CALIBRATION PROCE
DURE
Chapter
4
⑨
Alarm HI and LO limits
Overload of the ch1 value and ch2 value is monitored.
When they exceed the alarm HI limit or fall below the alarm LO limit, the external output "ALM"
turns ON.
Alarm HI limit
:
-19999
~
19999 (invalid at 19999 (initial value))
Alarm LO limit
:
-19999
~
19999 (invalid at -19999 (initial value))
* The setting value is common to ch1 and ch2.
・
For initial value, since both the alarm HI and LO limits are invalid as default settings, set them
again for use.
・
Since the alarm HI limit and LO limit are independent, simple use of either one of them is also
possible.
◇
Alarm HI and LO limits setting method
1.
Select setting mode 4.
Press
→
three times.
2.
Set "Alarm HI limit."
Press
seven times.
⇒
"Alarm HI limit" is
displayed.
Press
, and the highest digit blinks.
Input the setting value with
.
Move to the lower digits with
.
Validate it with
. (Initial value: 99999)
* For inputting a negative sign, press
while the input digit is blinking.
3.
Set "Alarm LO limit."
Press
once.
⇒
"Alarm LO limit" is displayed.
Press
, and the highest digit blinks.
Input the setting value with
.
Move to the lower digits with
.
Validate it with
. (Initial value: -99999)
* For inputting a negative sign, press
while the input digit is blinking.
4.
Press
twice to go back to the indicated value display.
Ch1 Ch
2
H I
LO
H I
LO
FNC
H I
Ch1 Ch
2
OK
LO
PEAK
HOLD
H I
LO
H I
LO
H I
Ch1 Ch
2
OK
LO
PEAK
HOLD
H I
LO
H I
LO
ESC